Exclamation My refractometer is way off when testing with Pinpoint 53mS calibration fluid

I got some Pinpoint 53mS salinity calibration fluid and tested my blue RHS 10-ATC (the cheap one everyone has) refractometer. It measured 1.030!!

The Pinpoint conversion chart shows it should read 1.0259, which is what I expect a calibration fluid would be set for. Now, if I check with plain RO, it does not measure 0, but is way below the scale.

I assume I should believe this calibration fluid?
__________________
[This space for rent]
  #2  
Old 10/18/2006, 08:02 PM
cwegescheide cwegescheide is offline
Moved out
 
Join Date: Oct 2003
Location: Tampa, Florida
Posts: 2,477
Randy has a recipie to make your own calibration fluid of 1.026 from a 2litre bottle and table salt. Worked perfect for me and verified it was accurate too.
  #3  
Old 10/18/2006, 08:24 PM
Billybeau1 Billybeau1 is offline
Team RC Member
 
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 6,369
Quote:
Originally posted by cwegescheide
Randy has a recipie to make your own calibration fluid of 1.026 from a 2litre bottle and table salt. Worked perfect for me and verified it was accurate too.
You must not have the cheapo.

Yes sjm, we have determined that the inexpensive refracts do not calibrate very well with RO or distilled.

My cheapo (properly calibrated) reads distilled water somewhere between - .0025 and -.0030

Here we go again
  #4  
Old 10/18/2006, 08:34 PM
cwegescheide cwegescheide is offline
Moved out
 
Join Date: Oct 2003
Location: Tampa, Florida
Posts: 2,477
Well brother pony up and get a NON-cheapo salinity meter!!! Get a nice refractometer and be done with it!!
  #5  
Old 10/18/2006, 08:37 PM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
I tried the DIY calibration fluid, but seemed to get a different result each time a made a batch. I guess my measuring cup skills are not that good. I never got the same amount of salt in the mix each time.

This is why I got the Pinpoint calibration fluid. Takes the me not making the calibration fluid error out of the equation.

So again, I can assume the Pinpoint fluid is accurate and now my refractometer is accurate?
__________________
[This space for rent]
  #6  
Old 10/18/2006, 08:38 PM
Billybeau1 Billybeau1 is offline
Team RC Member
 
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 6,369
Well brother, mine works just fine now that I know its properly calibrated.
  #7  
Old 10/18/2006, 08:39 PM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Quote:
Originally posted by cwegescheide
Well brother pony up and get a NON-cheapo salinity meter!!! Get a nice refractometer and be done with it!!
I can live with the one I have as long as it is properly calibrated, which I believe it now is.
__________________
[This space for rent]
  #8  
Old 10/18/2006, 08:42 PM
Billybeau1 Billybeau1 is offline
Team RC Member
 
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 6,369
Quote:
Originally posted by sjm817
Takes the me not making the calibration fluid error out of the equation.

So again, I can assume the Pinpoint fluid is accurate and now my refractometer is accurate?
Yes sjm, I've had trouble making the DIY solution too. I've spent so much time on this subject, my head still spins.

I'll get Boomer to explain it again. (I think)
  #9  
Old 10/18/2006, 08:54 PM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
I think $3.45 for a bottle of calibration fluid is money well spent, especially for the "measuring cup impaired" people like us.
__________________
[This space for rent]
  #10  
Old 10/18/2006, 08:58 PM
Billybeau1 Billybeau1 is offline
Team RC Member
 
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 6,369
Yes I agree. I just want to make sure your calculations are correct since this has been such an issue. Boomer will be along in a bit.
  #11  
Old 10/18/2006, 09:05 PM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
I just went from what the Pinpoint conversion chart shows. It shows the 53 mS solution = 1.0259

PINPOINT™ Salinity Monitor
Conversion Chart

mS | ppt | SG
51.0 33.50 1.0248
51.5 33.90 1.0250
52.0 34.30 1.0253
52.5 34.60 1.0256
53.0 35.00 1.0259
53.5 35.40 1.0262
54.0 35.70 1.0265
54.5 36.10 1.0268
55.0 36.50 1.0270
55.5 36.80 1.0273
56.0 37.20 1.0276
56.5 37.60 1.0279
57.0 38.00 1.0282
57.5 38.30 1.0285
58.0 38.70 1.0288
58.5 39.10 1.0291
59.0 39.50 1.0294
59.5 39.80 1.0296
60.0 40.20 1.0300
__________________
[This space for rent]
  #12  
Old 10/18/2006, 09:13 PM
Billybeau1 Billybeau1 is offline
Team RC Member
 
Join Date: Jul 2005
Location: Dyer, Indiana
Posts: 6,369
It doesnt surprise me that once you calibrated with the 53ms solution that your refract registers lower than 1.00 with ro or distilled. Thats my whole point.

My whole point (months ago) was that even though the manufacturer of the inexpensive ones says you can calibrate with RO or distilled, this is just not true.

I'm just not good as explaining the reasoning as Boomer.

My head still hurts.
  #13  
Old 10/19/2006, 12:55 AM
bertoni bertoni is offline
RC Mod
 
Join Date: Mar 2002
Location: Foster City, CA, USA
Posts: 35,743
The calibration solution for a conductivity meter likely won't work at all for a refractometer. The conductivity of a solution and its refractive index are not necessarily correlated if the ionic balance isn't fixed. So the conductivity solution is set up to have a certain "resistance", and its components might not at all resemble seawater, so its refractivity could vary greatly from seawater with that same "resistance".

Bottom line: you can't use a calibration standard for both a conductivity meter and a refractometer unless it was carefully designed to work with both.
__________________
Jonathan Bertoni

Last edited by bertoni; 10/19/2006 at 03:25 AM.
  #14  
Old 10/19/2006, 05:32 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Thanks for the reply. That makes sense. I'm going to send American Marine (Pinpoint) an email and see what they say too.
__________________
[This space for rent]
  #15  
Old 10/19/2006, 06:22 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Wow, I emailed AMI @ 6:30 this morning and got a reply @ 7:00! He said the calibration fluid should read 1.026 on a refractometer.
__________________
[This space for rent]
  #16  
Old 10/19/2006, 07:14 AM
Randy Holmes-Farley Randy Holmes-Farley is offline
Reef Chemist
 
Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 52,068
Hold on everyone!

The 53 mS standard provided by pinpoint may ONLY acceptable for conductivity meters. It is not likely to be 53 mS/cm seawater, but rather another chemical, and so will not necessarily match the refractive index of 53 mS/cm seawater and cannot necessarily be used to calibrate a refractometer.

That said, Lou Dell may have designed it to be suitable for a refractometer as well, but I am skeptical of that, and I'll invite him here to confirm this.
__________________
Randy Holmes-Farley
  #17  
Old 10/19/2006, 07:31 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
The email I got was from Lou Dell:

Quote:
It should read 1.026 on your refractometer. If you send me the bottle I
will verify it and send it back...Lou Dell/AMI
__________________
[This space for rent]
  #18  
Old 10/19/2006, 09:24 AM
Randy Holmes-Farley Randy Holmes-Farley is offline
Reef Chemist
 
Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 52,068
I know you said that, Lou emailed me this morning the same, but I am still skeptical and will continue to follow up with him.
__________________
Randy Holmes-Farley
  #19  
Old 10/19/2006, 09:59 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Thanks Randy. As usual, your help is greatly appreciated. I swear, I dont know how I would be able to keep my system going without help from experts such as yourself.
__________________
[This space for rent]
  #20  
Old 10/19/2006, 10:58 AM
Boomer Boomer is offline
Older Than the Cretaceous
 
Join Date: Aug 2002
Location: Duluth, Minnesota
Posts: 7,679
I'm going to have to see proof that a PinPoint 53 mS works on a refract. No, it is not impossible but I wonder if Lou is just confused. However, it would be great if it did work on a refract and EC meter or even a SG hydrometer. There are solutions that will work on all three but are very, very pricey.

I might add Lou's table is off also

53.0 35.00 1.0259

That table to be std, which it should be, should be

53.0 35.00 1.0264 and not 1.0259


That would have to be a table for a hydrometer calibtated @ 77F and being used in 80F water. And he shows no ref for temp for SG
__________________
If you See Me Running You Better Catch-Up


An explosion can be defined as a loud noise, accompanied by the sudden going away of things, from a place where they use to be.
  #21  
Old 10/19/2006, 11:47 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Randy, Boomer, Billy. If you would like, I can send some of the fluid to you and you can check it against your refractometer and analyze it. PM me if interested.
__________________
[This space for rent]
  #22  
Old 10/19/2006, 11:49 AM
Randy Holmes-Farley Randy Holmes-Farley is offline
Reef Chemist
 
Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 52,068
Well, I think we can work it out with Lou. He and I discuss things about his products reasonably frequently (and besides, I don't have a refractometer ).
__________________
Randy Holmes-Farley
  #23  
Old 10/19/2006, 11:52 AM
sjm817 sjm817 is offline
Generic Member
 
Join Date: Sep 2004
Location: Northern VA
Posts: 11,103
Quote:
Originally posted by Randy Holmes-Farley
Well, I think we can work it out with Lou. He and I discuss things about his products reasonably frequently (and besides, I don't have a refractometer ).
I guess you are still using a Hydrometer

I'm glad you guys interviened before I started raising the SG of my tanks. I'll hold off until things are confirmed.
__________________
[This space for rent]
  #24  
Old 10/19/2006, 12:08 PM
Boomer Boomer is offline
Older Than the Cretaceous
 
Join Date: Aug 2002
Location: Duluth, Minnesota
Posts: 7,679
We use very expensive EC meters, aka, Conductivity/Salinity meters These are the std for seawater.
__________________
If you See Me Running You Better Catch-Up


An explosion can be defined as a loud noise, accompanied by the sudden going away of things, from a place where they use to be.
  #25  
Old 10/19/2006, 12:18 PM
Randy Holmes-Farley Randy Holmes-Farley is offline
Reef Chemist
 
Join Date: Apr 2001
Location: Arlington, Massachusetts
Posts: 52,068
I guess you are still using a Hydrometer

No, I like conductivity.

FWIW, the Pinpoint conductivity probe is a good way to go, although I use a more sophisticated lab meter.
__________________
Randy Holmes-Farley
 


Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is On

Forum Jump


All times are GMT -5. The time now is 09:11 PM.


Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Use of this web site is subject to the terms and conditions described in the user agreement.
Reef Central™ Reef Central, LLC. Copyright ©1999-2009